Use of Adapters in Network Analysis
ANA Techniques where Calibration Kits are Not
Available
Measurements made with network analyzers are made with respect
to reference conditions established using a calibration kit in a
procedure known as "error correction". But what can be
done when presented with a device for measurement and yet a calibration
kit isn't available for that connector type? The usual course of
action is to use adapters and the measure at the adapter interface.
In many cases this method is adequate, although the measured values
include the effect of the adapters which must removed in some way
if less uncertainty is required. Many network analyzers have an
"adapter removal" routine built into the operating firmware
but, for this to work effectively, additional calibrations to characterize
the adapters are required. Consideration must be given to the trade-off
in accuracy against the time required to complete the measurements.
The first consideration is the acceptable quality of the measurement.
In terms of measurement uncertainty, the aim should be for potential
errors to contribute to the overall uncertainty budget in a reasonably
"insignificant" way. A case can be made (see article Single
Measurement Uncertainty) that "insignificant" can
be defined as up to, say, 30% of the total uncertainty. Performing
measurements using the "adapter removal" routines will
usually produce the best results, but the measurement and set-up
time will be much longer than the other methods that will be described
later.
Effect of Adapters
The accuracy of measurements performed without using the adapter
removal routine will depend on the quality of the adapters employed.
Corrections will normally be required whose precision will also
be influenced by the quality of the adapters.
The adapters chosen must enable the item to be measured to be connected
to the test port cables directly. Ideally, no more than one adapter
should be used on each lead to minimize the impact on measurement
quality although, in some cases, two "high quality" adapters
will produce smaller uncertainties than one "lower" quality
adapter. Follow the procedure shown above but note that in most
cases it is possible, if only correcting for the insertion loss
(and phase) of the adapters it is usually possible to save the adapter
measurements in the network analyzer's memory and then subtract
them from the measurement of the adapters plus the item. If high-grade
adapters are employed, it is often unnecessary to make corrections
for reflection magnitude -- usually just increasing the measurement
uncertainty (due to the uncorrected adapter) is adequate. Many high-grade
adapters have a reflection coefficient of better than 0.02 across
their operating frequency range.

Figure 1 -- Reflection
coefficient measurement of an APC7 0.15 rho air line following a
full, 2-port correction of the ANA using an APC7 cal kit. Adapters
are added and the measurement remade using the same calibration
constants. The difference between the red and yellow traces
represents the effect of the adapters. Removal of the air line and
connecting the adapters together provides the third trace.
Assumptions Ease the Problem
Adapters are easiest to measure in insertable pairs and it is best
to use adapters that are of similar physical length. This allows
an assumption that each contributes equally to the measured loss
and phase shift. The reflection coefficient of the adapter-pair
should be measured but can only be applied in corrections to the
data for the device-under-test (DUT) and adapter combination if
phase and magnitude measurements are made for all four S-parameters.
This can be time-consuming and a short-cut is just to use the adapters'
reflection coefficient magnitude as an uncorrected error contribution
in the uncertainty budget. A common assumption is that adapter reflection
coefficient magnitudes add in a quadratic fashion.

Figure 2 --Reflection
phase measurements on the same port of the 0.15 rho air line.
Again, the measured reflection phase of the adapters has been subtracted
from the measurements made on the air line with adapters. Essentially,
the two traces overlay each other so the third trace shows the phase
difference of less than 0.5 degrees. Such small phase differences
are generally difficult to discern as the stability of cables and
system repeatability often exceeds this value.
Non-insertable devices pose some of the most difficult measurement
problems, even when using the adapter removal capability of modern
network analyzers Again, it is important to understand customer
requirements as it might be acceptable to perform the ANA error
correction in a convenient connector-type and then measure the DUT
with an adapter. The measured values include the adapters but most
precision-grade adapters have a fairly negligible effect at low
to medium microwave frequencies. In cases where better accuracy
is desired, the adapter characteristics must be determined. For
non-insertable items with similar connector-type "X",
it may be possible to use adapters that are sold in "phase
and loss" matched pairs. These work on the principle that the
X-to-male adapter has the same physical loss and phase shift as
the X-to-female adapter so that the loss of two X-to-male adapters
can approximated by measuring the insertable pair.

Figure 3 -- Transmission
(loss) magnitude measurement of the 0.15 rho air line. The effect
of the adapters has been removed by simple subtraction. This technique
appears to work well where the match of the device approaches 50
ohms (the bottom of the peaks) but as the match worsens (peak top)
the difference between using or not using adapters becomes more
obvious. Also note that the effect of mismatch loss uncertainty
does not always produce higher loss; some combinations have produced
lower measured values than might be expected.
Further Difficulties
Unfortunately, some assumptions are made which are difficult to
prove. Many manufacturers make male and female adapters that they
advertise as being of similar loss and phase characteristics. This
might be "proved" by assuming that the reflection parameters
are similar. The approach relies on these similarities to arrive
at an assumed loss and phase shift value for each adapter. The accuracy
of this technique can be improved slightly by measuring each adapter
with a similar one of the opposite sex (to form an insertable pair)
and using the average measured values as corrections. At the expense
of increased measurement and calculation time, even better confidence
can result from making at least three measurements using different
adapters and solving simultaneous equations to determine each adapter's
characteristics.
Air lines have a predictable response but, since they have little
transmission loss, imperfections in the measurement equipment, adapters
or cables can be imposed on all measurements. For an attenuator
pad, the effects of mismatch loss uncertainty and the uncertainties
in the measurement of the mismatch tend to be isolated to the parameter
being measured. For example, the mismatch measurement on port 1
will be influenced by the termination of port 2 to a much lesser
degree. Therefore, the isolation provided by an attenuator more
readily enables investigation of the individual parameters.

Figure 4 -- This plot
shows reflection magnitude measurements made on a 10dB APC7 attenuator.
The effect of the adapters is clearly seen and because the adapters'
mismatch is similar to the attenuator, the additional error becomes
quite significant. It is interesting to try to correlate the plots
but it's not immediately obvious as only the magnitude term is shown.
The errors are generated as the phase vectors rotate, causing the
effects of mismatch to interact with each other. The worst deviation
might therefore occur away from the peaks of mismatch, but where
the phase components have also combined to greatest effect.

Figure 5 -- The transmission
magnitude measurement using adapters looks much more predictable.
Notice again that although we refer to "mismatch loss uncertainty",
the term can appear to have "gain" (i.e. in the 2GHz to
6GHz region of the plot). The attenuator has a reasonably good match
at its ports, as do the adapters, so the apparent error when using
the adapters is, in this case, quite small.
Non-insertable devices with dissimilar connectors are the most
difficult to measure. Devices often falling into this category are
couplers, splitters, combiners and sometimes cables. The method
is to measure each adapter with a similar one of the opposite sex.
As before, common practice is to ignore the effects of the adapters
on the DUT reflection measurements, instead making additional allowance
in the uncertainty budget. An approximation for this additional
contribution can be obtained during the measurement of the adapters'
insertion loss by setting the ANA to also display S11
or S22; each adapter contributing in a quadratic manner.

Figure 6 -- Here, the
transmission phase of the 10dB attenuator is measured. With some
isolation between the ports and reasonably good matches, phase measurement
works well even with adapters. The differences shown are well within
the repeatability of the analyzer and its cables.
The Time Domain Solution?
Time domain gating is a process whereby the measurements made in
the frequency domain are mathematically converted into the time
domain. Any discontinuities in the transmission line (such as connections)
then show as a spike on the analyzer's display and can be "gated
out". The measurements can then be converted back into the
frequency domain and the characteristics of just the DUT displayed.
But time domain measurements can be problematic. In order to obtain
the best resolution, high frequencies must be used but, unfortunately,
measurements in the upper and lower 10% regions of the frequency
band cannot be considered reliable. The Time Domain function on
any network analyzer requires a lot of skill to use; it is easy
to make repeated measurements but get very different results each
time.
Summary
- Reflection coefficient of the adapter(s)
often dominates the measurement uncertainties.
- The adapter(s) insertion loss and phase
shift can be readily corrected for and will tend to have little
effect upon the measurement uncertainties.
- Lower grade adapters are more likely
to have reflection coefficients that are sufficient to degrade
the equivalent load and source match terms of the network analyzer
that consequently impacts the accuracy of the adapter insertion
loss and DUT attenuation measurements.
|