Traceability of Vector Network Analyzers
About this Article
Written by Ken Wong of Agilent's Santa Rosa
manufacturing division, this article originally appeared in Test
& Measurement Europe magazine, February-March 1996. Copyright
1996 Reed Elsevier, Inc. and used with permission.
Introduction
Measurement traceability is becoming an international requirement
with the world wide acceptance of ISO quality standards. The issue
of traceability, however, is often confusing, especially for microwave
network analyzers. There are many definitions of traceability. It
is not obvious how the measurement parameters are linked to the
basic fundamental constants such as frequency, volt, length etc.
The objective of this paper is to clarify these issues.
Traceability must be defined in conjunction with calibration, verification,
and the associated uncertainty of a measurement system. The internationally
accepted definitions of these terms is documented in ISO Guide 25,
General Requirements for the Competence of Calibration and Testing
Laboratories. The intent of paragraph 9 of ISO Guide 25, "measurement
traceability and calibration", can be summarized as follows:
"Traceability to designated standards (national, international,
or well characterized reference standards based upon fundamental
constants of nature) is an attribute of some measurements. Measurements
have traceability to the designated standards if and only if scientifically
rigorous evidence is produced on a continuing basis to show that
the measurement process is producing measurement results for which
the total measurement uncertainty relative to national or other
designated standards is quantified."1
VNA Calibration
Vector network analyzers (VNAs) measure the relationship between
incident waves and reflected and transmitted waves. Measured quantities,
such as coefficient of reflection, return loss, insertion loss,
forward gain, and S-parameters are referenced to the defined impedance
of the measurement system.
To minimize measurement uncertainties, VNAs are calibrated using
a set of "known " impedance calibration standards such
as opens, shorts and loads. The measurement uncertainty of the calibrated
VNA system is directly proportional to the uncertainty of the known
quantities of these standards. The impact of each standard's uncertainty
depends on the calibration method used, stability of the measurement
system, and uncorrected performance of the system as illustrated
in Figure 1.

Figure 1 -- Propagation
of calibration errors
Uncertainty of calibration standards
The key factors that affect the accuracy of RF/microwave calibration
standards are connector repeatability, dimensional accuracy, and
electrical characteristic accuracy. The most accurate microwave
calibration standards are metrology grade precision transmission
lines and shorts. (Metrology grade connectors are known also as
laboratory precision connectors, LPC.) The impedance and other microwave
characteristics of these devices are calculable from their physical
dimensions and physical properties of the material used.2
The uncertainty of these devices is a function of the physical
metrology systems.
Metrology grade devices are fabricated to the tightest specifications
possible using state-of-the-art equipment and methodology. At HP,
precision coaxial airlines can be fabricated with a diameter tolerance
of <2µm and diameter consistency of <1.5µm. This is equivalent
to an impedance accuracy of 0.2% for 7 mm devices and 0.6% for 2.4
mm devices.
For coaxial devices, the female contact must be slotless to minimize
performance dependency on mating devices, a prerequisite of metrology,
grade coaxial connectors.3 Metrology
grade standards reduce calibration uncertainties by a least 50%
over instrument grade standards. HP is the only source, worldwide,
for microwave metrology grade coaxial connectors. The electrical
characteristics of other calibration devices can be derived from
these physically calculable standards. These derived electrical
characteristics are then fitted to a model using calibration constants.
HP's calibration standards are machined to very tight tolerances,
<2.5µm for critical dimensions, and thus are interchangeable.
It is not necessary to keep track of which open or short to match
to which set of calibration constants or calibration model. Our
calibration constants are derived from very accurate computer modeling
and experimental data. The model accounts for skin loss, impact
of pin depth, and other imperfections.
Traceability of calibration standards
All HP standards are traceable, through precision airlines and
shorts and mechanical gages, to the National Institute of Standards
and Technology (NIST) in Gaithersburg, MD (see Figure 2).
These are the most precise standards available for network analyzers.
HP's traceable path is the same path that many national laboratories,
such as the National Physical Laboratory (NPL) of UK, are using.
Figure 2 also compares HP's traceable path with NIST's six-port
traceable path.

Figure 2 -- VNA traceability
and uncertainty
The measurement uncertainties of all the applicable mechanical
measurement systems are vigorously quantified using well established
statistical methods. An accurate electrical model, based on accepted
microwave theory and practices and verified data, is created to
convert the mechanical/physical properties into microwave S-parameters.
The uncertainties also are allowed to propagate through the S-parameter
calculations and accounted for in the uncertainties of the S-parameters,
as diagrammed in Figure 3.

Figure 3 -- Uncertainty
propagation
The modeled S-parameters are then used to calibrate HP's VNA measurement
system. This calibration method removes any imperfections that the
airlines and shorts may have, and provides the link to the fundamental
units of measure length and other physical constants.
VNA verification
To ensure that the VNAs measurements are accurate as specified
after calibration, the measurement system must be verified. An internationally
accepted method of VNA verification uses a set of verification standards,
such as a low loss match airline, a mismatched device and two lossy
devices. These devices are first measured by a reference laboratory.
The measured data from the VNA in question is compared with the
data obtained from the reference laboratory. If the two measurements
are statistically similar, then the system is considered verified.
This method is also known as "tops down" verification.
Important considerations for selecting verification standards are
that :
- the devices must be physically different from the calibration
devices; and
- the devices cover an adequate range of reflection and transmission
characteristics .
The "tops down" verification method works best when the
measurement uncertainty of the reference laboratory is much smaller
than the measurement uncertainties of the VNA system in question.
Unfortunately, the best verification standards and the best calibration
standards are at the same accuracy level. Therefore, it is not possible
to verify a VNA system's accuracy by measuring a set of verification
devices using the traditional tops down method. In addition to performing
a tops down verification, statistical data of the verification results
should be kept to establish statistical system performance limits.
Participation in national and international "measurement comparison
programs" will provide added assurance of the VNA's performance
and proof of traceability.
A "bottoms up" verification approach has been proposed
and used by many calibration laboratories. This approach uses precision
airlines, and other physical standards in a different calibration
method to extract the "residual errors" of the original
calibration. Again, since the original calibration used similarly
derived standards, measurement comparisons of interlaboratory standards
are still needed. Other error sources of the measurement system
are also individually quantified at the most fundamental level.
This method provides a more thorough treatment of the VNA system
verification, but is much more time consuming and costly.
Conclusion
VNA systems are traceable through calibration and verification
standards. The quality and accuracy of these standards determine
the achievable accuracy of a VNA . No calibration method can overcome
this fundamental limitation. HP provides the highest grade coaxial
and waveguide calibration and verification standards today. The
manufacturing process, metrology systems and device modeling methodology
used to ensure the integrity and accuracy of HP's standards are
closely and continuously monitored and improved so that our VNA
systems' measurement accuracy are solidly supported.
References
- Belanger, B.C., "Traceability: An Evolving
Concept", ASTM Standardization News , Vol 8, No. 1 1980,
pp22-28.
- Wong, K.H., "Characterization of Calibration
Standards by Physical Measurements", ARFTG Digest Spring
1992.
- IEEE Standard 287, "Precision Coaxial Connectors",
(draft version).
|