Calibration, Verification or Conformance ?
When discussing calibration requirements with
a potential supplier it's obviously important to understand what's
being offered. Other articles in this section should help you to
establish your requirements and distinguish the differences between
available services. But one of the variations has, sometimes, even
confused calibration laboratories and quality auditor. It's
a matter of the difference between calibration, verification and
even conformance.
Similar to the often confused specification terms accuracy
and precision, a myth became "established wisdom"
that calibration and verification are differentiated on the basis
of quality or integrity.
Popular opinion being that verification is a quick-check of performance
perhaps made without any real traceability, whereas calibration
provides genuine assurance that the product really meets its specification.
In fact, the US national standard ANSI/NCSL-Z540 defines "verification"
as being calibration and evaluation of conformity against a specification.
This definition originated with the now obsolete ISO/IEC Guide 25
but neither its replacement (ISO/IEC 17025) or the International
Vocabulary of Measurement (VIM) currently have it or any alternative.
The only relevant international standard that includes terminology
covering the process of both calibrating and evaluating a
measuring instrument's performance against established criteria
is ISO10012 which uses the rather cumbersome term "metrological
confirmation".
Calibration is simply the process of comparing
the unknown with a reference standard and reporting the results.
For example:
Applied= 1.30V, Indicated= 1.26V (or Error= -0.04V)
Calibration may include adjustment to correct any deviation
from the value of the standard.
Verification, as it relates to
calibration, is the comparison of the results against a specification,
usually the manufacturer's published performance figures for the
product. (e.g. Error= -0.04V, Spec= ±0.03V, "FAIL").
Some cal labs include a spec status statement on their Certificate
of Calibration. (i.e. the item did/did not comply with a particular
spec).
Where no judgment is made about compliance, or correction has not
been made to minimize error, it has been suggested that Certificate
of Measurement would be a more descriptive title to aid recognition
of the service actually performed. Some suppliers also use Certificate
of Verification where no measurements are involved in the performance
testing (such as for certain datacomm/protocol analyzers), rather
than Certificate of Functional Test as this latter term
is often perceived as simply being brief, informal checks as might
be performed following a repair (often termed "operational
verification").
Verification can also relate to a similar evaluation
process carried out by the equipment user/owner where the calibration
data are compared to allowances made in the user's uncertainty budget
(e.g. for drift/stability between cals) or other criteria such as
a regulation or standard peculiar to the user's own test application.
Verification is not intermediate self-checking
between calibrations. Such checks are better termed confidence
checks, which may also be part of a Statistical Process Control
regime. The results of confidence checks may be used to redefine
when a "proper" calibration is required or may prompt
modification of the item's working spec as assigned by the user.
But what about conformance, especially regarding
the meaning of a Certificate of Conformance ? Typically
available when an instrument is purchased, it is now generally
recognized that such a document has little value as an assurance
of product performance. Of course, the manufacturer expects that
the product conforms to its spec but, in this sense, the document
simply affirms that the customer's purchase order/contract requirement
has been duly fulfilled.
|