Accreditation for Complex Electronic Instruments

About this Article

This article, written by Dave Abell of Agilent Technologies, Santa Clara, was presented at the May 2001 Simposio de Metrología by Sergio Lopez-Carmona. The conference was organized by the Mexican national metrology institute, CENAM.


Summary

Accreditation for calibration has been long sought after by suppliers and users as a way to reduce multiple independent audits. With the adoption of ISO/IEC 17025 and the rapid development of mutual recognition through organizations such as ILAC and NACLA, the groundwork has been laid for a robust, internationally recognized system for calibration laboratory accreditation. As the old axiom goes, caveat emptor, "let the buyer beware". Purchasing calibrations from an accredited laboratory for a complex electronic instrument doesn't assure that you will necessarily receive a better or even more complete calibration than from a non-accredited laboratory. This presentation explores the questions to ask when having an instrument like a spectrum analyzer or precision digital multimeter calibrated.

Choosing the Right Standard

Several standards address the calibration of test equipment, but one has achieved almost universal adoption as the basis for accreditation: ISO/IEC 17025. The general quality standard, ISO 9000:1994, has a section titled 4.11 Inspection, Measuring, and Test Equipment that lays out the basic requirements for calibration. Section 4.11 also points out the need for adequate measurement uncertainties in the related procedures. Typically an assessor to ISO 9000 doesn't have the technical background to make a judgement on the adequacy of calibration procedures. ISO 9000:2000 7.6 Control of Measuring and Monitoring Devices references ISO 10012 for guidance, but this standard is not normally utilized by laboratory accreditation agencies.

The use of an accredited laboratory is meant to reduce multiple audits to a supplier. Accreditation, unlike registration of a quality system, adds the dimension of technical proficiency to the assessment. An accredited laboratory has been evaluated not only on the robustness of their quality system, but also their technical proficiency to determine if they can actually make the measurements to the degree of accuracy claimed. If a laboratory is accredited in this manner, it eliminates the need for second party audits since the purchaser can rely on the results from the third party accreditation assessor.

It is for this reason - technical proficiency and assurance - that some industries and customers rely on accreditation as the hallmark of a capable laboratory. QS9000 developed by the big three automotive companies - Ford, General Motors and Daimler Chrysler - is moving to require accredited calibrations for suppliers. One of the difficulties they are encountering, however, is the scarcity of accredited labs for calibration. Until there are sufficient laboratories available, the Automotive Industry Action Group (AIAG) recommends some alternatives such as second party audits. See the AIAG web site for details.

ISO 17025 General requirements for the competence of testing and calibration laboratories applies both to companies that calibrate testing equipment as well as those that do materials and other kinds of testing. Also, ISO 17025 is based on ISO 9000:1994, not ISO 9000:2000. This may require some suppliers to carry dual elements in their quality systems - one for the '94 version and one for the '00 release of ISO 9000.

Calibration Quality System Basics

Figure 1 shows at a conceptual level the components of a quality system required by the ISO 17025 standard. The assessment done for accreditation to ISO 17025 only reviews the calibration operation of a business. Other services, such as repair provided by a calibration laboratory, are not included in the scope of ISO 17025 assessment. 17025 review is not a substitute for the broader evaluation done by an ISO 9000 auditor. The elements of the quality system are essentially the same, however, and there is a great deal of leverage for a supplier between these two quality systems.

Figure 1 -- Basic requirements of ISO/IEC17025

figure 1

The unique difference for the evaluation by ISO 17025 is the assessment of the technical capability of the laboratory to perform the work required by the customer. For the calibration of electronic equipment, this is highly dependent on the laboratory's traceability and measurement uncertainty as indicated by stars in Figure 1. As defined by the International Vocabulary of Basic and General Terms in Metrology (VIM), traceability is the "property of the result of a measurement or the value of a standard whereby it can be related to stated references, usually national or international standards, through an unbroken chain of comparisons all having stated uncertainties." Traceability assures that the measurements made by the laboratory are comparable to those made by others, a fundamental requirement for fairness in trade. Knowing measurement uncertainties associated with the procedures allows an assessor to judge the adequacy of the laboratory's calibration methods.

Checklist for Calibration Services

Philip Stein in "Calibration Buyers, Beware!: What to look for when considering a commercial calibration laboratory" (American Society for Quality's Quality Progress magazine, Sept. 2000) suggests these questions be addressed:

  1. Is the laboratory …accredited?
  2. Is the body that accredited this laboratory a signatory to one of the lab accreditation agreements?
  3. Are the measurement parameters you wish to have calibrated listed on the laboratory's scope of accreditation? Are the ranges of the parameters you have chosen within the scope?
  4. Have you specified accredited service on purchase order to the laboratory?
  5. Do all the certificates you received from the laboratory have a logo from the accreditation body, and are no exceptions taken for specific results?

This is good advice, especially item three. If the device to be calibrated is simple, such as a gage block, then it is straight forward to compare the parameters to be calibrated against the scope of the laboratory's capability. In the case of a complex electronic instrument, this comparison may not be so simple, however.

Scope versus Equipment Specifications

Most users of an electronic instrument simply want to know if it meets the specifications as advertised by the manufacturer. The accreditation systems in use today are the product of measurement professionals at various national laboratories, with minimal input from manufacturers of electronic instruments. Calibration services provided by national laboratories don't reference an artifact's specifications, but provide a detailed and precise report on the characteristics of the device being tested. This is one of the problems with calibrations for electronic test equipment -- the accreditation system is based upon providing an objective measure of characteristics, not stating that the device operates within a series of specifications.

Accredited calibration laboratories can, of course, provide a calibration within their scope of capabilities that reports whether or not an electronic instrument meets its specification. The challenge for the purchaser of calibration services is to choose the laboratory which has the necessary capabilities to provide a statement of conformance to specifications. An instrument specification is typically written by a manufacturer in a different form than the way in which an accrediting agency writes a scope of accreditation. This means that the purchaser must make a translation between these terms and formats to determine the adequacy of a laboratory's capabilities. The accredited calibration laboratory may have already done this if it has calibrated the specific model in question before. If not, the laboratory may have to do some research before it can state whether or not it can fully calibrate an instrument to manufacturer's specifications.

Figure 2 -- Scope versus equipment specification

figure 2

Specifications for an electronic instrument are derived from the design characteristics of the instrument (see Figure 2). In many cases, the instrument is not actually 'calibrated' on a test system at the end of the line, but the specifications are assured through the manufacturing process. Constant measurement of this process makes it possible to be confident that the instrument fully meets its specification when it reaches the end of the production line. The specifications may not be directly tested, but inferred from other operating parameters of the instrument. Users also like to see a number of 'typical' specifications to better understand the operating characteristics of the instrument. These typical specifications are usually not guaranteed, but given as an example of nominal performance. Reputable companies will stand behind their products, but matching up an obscure 'typical' specification with a scope of accreditation may be difficult.

The scope of accreditation is written for a specific International System (SI) unit at a particular level of measurement, with its corresponding uncertainty. The manufacturer's calibration procedure is written to insure all of the advertised specifications are within the limits called out in the data sheet. In some cases, the actual capability of the instrument may exceed those specified in the data sheet. The best combination is having the scope of accreditation cover all of the stated specifications with the measurement uncertainty needed to assure the data sheeted performance. For simple instruments such as DC voltmeters, this may be possible. For more complex instruments such as microwave signal analyzers, this may be very difficult to accomplish.

The purchaser of calibration services should also be aware that assigned uncertainties by the laboratory may be larger than the published scope and may also be larger than the specification precluding the inclusion of a compliance statement. The laboratory's scope for an RF power measurement may have been done using type-N connectors, but the instrument in question uses a different kind. In this case, the recipient of the calibration report must determine if the measured characteristics are adequate for the intended use. To further complicate the matter, an instrument specification may be related to SI units through more than one parameter. Resolution bandwidth is one such example.

An experienced lab should communicate clearly to a customer what they can and cannot provide in the way of service. An informed customer will ask the question "can you provide a fully accredited calibration on this instrument?" If the answer is 'no' the lab must tell which specifications must be eliminated or done with greater measurement uncertainty.

Three fundamental questions should be asked when determining if an instrument can be fully calibrated:

  • Is the range of parameters available?
  • Is the required cal method available?
  • Is the measurement uncertainty of the laboratory's accredited scope sufficient?
Simple example: DC Voltage

The top table in Figure 3, is a excerpt from the data sheet specifications for the instrument. The bottom table is an excerpt from the scope of accreditation for a laboratory when generating DC volts. Notice that the data sheet specifications are in '% of reading + % of range' whereas the scope document is for generating a voltage with 95% confidence '6 ppm + 0.4uV' at up to 0.22V. The scope also includes the equipment that is used to meet this specification.

Figure 3 -- Comparing specifications for an Agilent 34401A voltmeter
against a calibration laboratory's accredited best measurement uncertainties

figure 3

Without any other information, we must assume that the instrument specification is at least 95% confidence (k=2). Many electronic instruments were designed before the wide acceptance of the Guide to Expression of Uncertainty in Measurement (GUM), so the design teams may not have used the specific GUM language and methodology. Most instrument manufacturers are, however, conservative in nature and it is safe generally to assume at least 95% confidence.

In this case, the comparison between the manufacturer's specifications and the laboratory's scope of accreditation is quite simple, requiring only one intermediate step (see Figure 4).

Figure 4

figure 4

First, choose the voltmeter's data sheet table for one year specifications at 100mV. The accuracy as a percent of reading and as a percent of range is listed at 0.0050 and 0.0035 respectively. Using simple multiplication, this means that the indicated voltmeter reading is ± (5 mV + 3.5mV) or a total of ± 8.5mV maximum. Looking at the laboratory accreditation scope, 100mV is covered by the laboratory's capability at the 0.22V range. The measurement is listed as the ability to generate a voltage with a specific accuracy at this range. In this case, the accuracy is listed in ppm, or 'parts per million' which is 1/1000000 or 0.0001%. Thus, 100mV generated on this range has an accuracy of ±(6ppm*100mV + 0.4mV ) or ±(0.6mV + 0.4mV ) or a maximum total of ±1.0mV . This is well within the specification of the voltmeter. As long as the laboratory uses the method that was prescribed by the assessor for this voltage range, the specification of the voltmeter can be assured.

Complex example: Resolution Bandwidth

In the next example there is not a one-to-one correspondence between the instrument specification and the scope of accreditation. Figure 5 is a section of the catalogue specifications for a series of spectrum analyzers made by Agilent Technologies which extend up to 26.5GHz.

Figure 5 -- Extract from a typical spectrum analyzer specification

figure 5

One of the first specifications is Resolution Bandwidth (RBW). For the Agilent ESA series of signal (spectrum) analysers, the accuracy is stated as ±15% for 1kHz to 3MHz and ±30% for 5MHz settings. What does it mean in relation to the typical scope of best measurement capability published by an accredited laboratory? RBW is an example where the specified function doesn't appear in the capability listing, although the performance is stated in terms that are listed such as Hz or frequency.

Is the uncertainty for this test defined by the lab's frequency capability? No, unfortunately it is more complicated than that. To determine the uncertainty it's necessary to investigate the test procedure. This may vary from lab-to-lab or even model-to-model. The RBW uncertainty is actually a function of amplitude accuracy (y-axis) as attenuation or power and knowledge of the input filter shape, specifically its roll-off characteristic, translated into equivalent frequency (x-axis).

Figure 6 -- Relationship between amplitude and frequency uncertainty

figure 6

The diagram in Figure 6 shows the frequency uncertainty only on the left in the expanded view. Essentially the same figure must be added for the right side of the signal curve. The typical uncertainty for this product is 1% -- nothing like the frequency uncertainties stated in the schedule that might initially be thought to relate to the RBW specification. For example, the accreditation schedule for an Agilent Technologies laboratory in the UK states frequency accuracy of 2 x 10-10 to 4 x 10-9 in the 0.01 Hz to 100 MHz range. This is considerably smaller than the actual capability for RBW at 1%. The exact calculation of accuracy available from the laboratory depends upon the error equation of the method used, which includes both frequency accuracy and RF power amplitude accuracy.

The customer alone can't determine the extent of calibration or the possible uncertainties from the published accreditation scope schedule. They will need to rely on the deeper analysis done by the supplying laboratory. This may also cause delays in ordering calibration since the laboratory may not have calculated the error equations for the specific model being requested, and may have to do considerable research.

Again, the most direct route is for the customer to ask if the laboratory has ever calibrated this model instrument before and if so, what limitations arose.

Conclusion

All accredited calibrations are not equal. A laboratory could have a simple scope of accreditation for DC/AC volts based upon a single commercial calibrator, or an extensive scope including arcane procedures for difficult optical or microwave parameters. The buyer must be aware of the parameters that are critical for their use when choosing a laboratory.

In any case, a buyer must be flexible when purchasing accredited calibrations. Even the best of laboratories may not have the accredited capabilities in optical power, phase noise measurements or other state of the art parameters. In fact, some leading edge instruments are difficult to calibrate, and some can only be done to full accuracy at the manufacturer's production facility.

However, there is value in purchasing from an accredited laboratory. Even if their accredited uncertainties are too large to verify an instrument's performance against its published specification, the fact that they have undergone a rigorous evaluation emphasizes their commitment to quality and technical proficiency of their measurement services.

Like most transactions, it is important to establish a level of trust and evaluate the reputation of the calibration provider. It also is good to work with a laboratory that regularly calibrates the type of complex electronic instrument you own since they will more likely have done the detailed analysis need to connect the instrument specifications with the scope of accreditation. Accreditation goes a long way towards improving this selection process but, one must always keep in mind, however, it is not a comprehensive solution in this measurement arena.

Acknowledgements

Michael Hutchins of Agilent Technologies, United Kingdom contributed the example for resolution bandwidth. Greg Burnett of Agilent Technologies, United States contributed the example for DC volts. Both helped greatly with ideas and proofreading.

Bibliography
  1. Examples of Schedules of Accreditation
  2. Hysert, Gary, The CLAS Way to Cal Lab Accreditation (Boulder Colorado: NCSL International Proceedings NCSL 2000).
  3. Stein, Philip. Calibration Buyers, Beware !
  4. Automotive Industry Action Group AIAG, QS9000 requirements
  5. International Vocabulary of Basic and General Terms in Metrology (Geneva, Switzerland International Organization for Standardization, 1993).
  6. ISO/IEC 17025. General requirements for the competence of testing and calibration laboratories (Geneva, Switzerland International Organization for Standardization, 1999).