Instrument Design Validation and Recommended
Calibration Policy
Introduction
What are Agilent's policies regarding the design
of the recommended performance test published in our Service Manuals?
Why should users have confidence in the overall performance of the
instrument, even for functions and ranges that don't seem to be
included in the calibration procedure? Although the following article
addresses these questions from the perspective of the manufacturing
division responsible for digital multimeters, the general principles
also apply to other product-types.
Digital
Multimeter Adjustment and Verification Procedures
From time to time, customers or calibration laboratories may inquire
about why specific adjustment and verification points or procedures
are selected for a given instrument model. This discussion is intended
to provide general background information with respect to the methods
and philosophies that Agilent Technologies utilizes when specifying
these aspects of individual product service procedures.
Calibration and verification procedures documented in the Agilent
Service Guide are created and reviewed by design, service, and quality
engineers and incorporate our detailed, proprietary knowledge of
the DMMs internal hardware and software design and sources
of measurement error. Development of procedures and selection of
verification test points is based upon our extensive statistical
analysis of both characterization data gathered during design verification
testing and through on-going monitoring of production processes.
During Agilent manufacturing, significantly more verification data
are gathered and used to monitor product performance and to assure
our outgoing product quality. The documented user procedures completely
describe all steps required to fully adjust an instrument to conform
to its published accuracy specifications. Philosophically, Agilent
verification procedures are designed to achieve >99% confidence
that the instrument conforms to all published measurement specifications
and that it is fully functional for use. This high level of user
verification confidence is achieved by a multi-tiered approach as
described below.
First, all accuracy verification procedures are preceded by checking
basic operational readiness through executing the instruments internal
Self-Test procedure. This checks internal circuit paths for functional
operation and is intended to assure, with >90% confidence, that
the instrument has not experienced a hardware failure and should
be expected to meet all published measurement specifications
if the specified adjustment procedures have been followed
previously. Some non-measurement, user accessible, functionality
(e.g. display, keyboard, computer interface, etc.) can not be completely
verified by Self-Test and are generally not addressed by measurement
verification procedures. Certain instrument models utilize internal,
Auto-calibration procedures that should be executed before any performance
verification checks are performed. Auto-calibration, when employed,
automatically compensates for numerous measurement gain and offset
drifts due to operating temperature variation and component aging
effects. Auto-calibration utilizes internal transfer measurements,
relative to the instruments primary voltage and resistance reference
standards, to eliminate these measurement errors. Second, all zero
offset calibration points are verified including both front and
rear input terminals, where present, since separate offset values
are stored for each during adjustment.
The third tier of performance verification confidence comes from
verifying the linear gain terms of each unique measurement path.
For example, while two-wire ohms and 4-wire ohms appear to the user
as two independent measuring functions, they in fact share near
100% of the same measuring circuits, differing only in the offset
portion of the measurement. The ohms current source, responsible
for the linear gain term of the measurement is shared in both functions.
Generically, gain verification is performed near the full range
points using the nearest commonly available value. For example,
ohms full range values are in multiples of 1.2 (e.g. 120W,
1.2kW, 12kW,
etc.) while Agilent specified adjustment and verification values
are chosen in standard multiples of 1.0 (e.g. 100W,
1kW, 10kW,
etc.) for ease of user support. The verification test points and
methods specified by Agilent are selected to achieve maximum performance
verification confidence while not requiring undue support or cost
of ownership burden on our customers.
The fourth element of the verification procedures is aimed at
validating the performance of other circuit paths not specifically
addressed by the linear offset and gain terms previously discussed.
For example, this includes verification checks of the analog-to-digital
converter (ADC) linearity (guaranteed by design and not adjusted)
and of the ac signal conditioning path frequency response which
may be either wholly guaranteed by design or may be adjusted at
a single cross-over frequency. Since the same ADC is employed for
all measuring functions and ranges, its characteristic is verified
in a single configuration where the signal conditioning circuits
have the least effect on the overall measurement result. The ADC
integral linearity characteristic can be verified using several
measurements across the complete scale (i.e. positive full scale
to negative full scale). Similarly, the frequency response of the
ac section can be verified at the accuracy band edges on a subset
of the measuring ranges based upon specific knowledge of the instruments
circuit topologies. In addition, some ac measuring characteristics
are determined by fixed, digital signal processing algorithms (DSP)
and therefore do not require user verification. These behaviors
have been verified earlier through extensive product design validation
testing.
In summary, modern instruments such as DMMs employ closed-box
electronic calibration methods to store and digitally process measurement
correction constants for linear error terms. High quality instrument
designs minimize non-linear error terms by design such that no user
corrections are necessary to compensate for these non-ideal behaviors.
In addition, many traditionally analog behaviors of instruments
have been replaced by digital circuits, software algorithms and
digital signal processing techniques whose characteristics do not
change with time, temperature, etc. Therefore, many of the historical
beliefs and experiences of users and calibration laboratories that
developed with past generations of measuring instruments are becoming
increasingly obsolete and outdated; particularly when inferring
sources of measurement error in modern instrument designs. Since
independent verification of every possible measured value is, and
always will be, impractical by end users, one must rely on the guidance
and integrity of the instrument manufacturer to specify appropriate
adjustment and verification procedures for the instrument given
their detailed knowledge
of design limitations and instrument failure modes. As always, users
may augment the manufacturers verification procedures, as
they deem necessary, to achieve higher verification confidence at
application-critical measurement points.
|