Ratio Transformer (Inductive Voltage Divider)
Calibration
About this Article
This article by Toshiaki Aoki and Katsumi
Yoko was presented at the 1995 NCSL Symposium was published in the
Proceedings.
Abstract
This paper describes a technique for calibration of a decade ratio
transformer, or inductive voltage divider (IVD), which is performed
by comparison measurement at 1kHz. Comparison measurement is made
between a device under test and a standard ratio transformer which
is calibrated for the first decade by Japan Electric Meters Inspection
Corporation. A method using this approach is outlined that uses
a commercially available precision capacitance bridge, a four-terminal-pair
LCR meter, and three-terminal capacitors. Experimental results are
given to demonstrate the validity of the calibration technique.
View or Download
The entire paper is available to view or
download in Portable Document Format. But if your browser doesn't
already have the (free) viewer installed, you will first need to
get and install the Acrobat® Reader.
IVD
Calibration
|
|