Mind the
Gap !
Travelers
on London's "tube" (underground rail network) are
familiar with the warning that's announced at some stations
about the gap between the curved platform and train. It's
also a surprisingly valid consideration if you're trying to
make accurate microwave impedance measurements !
|
 |
Connector Pin Recession and its Effect on Network
Analyzer Accuracy
The apparent effect of a recessed connector upon network analyzer
measurements is often talked about, but without any quantitative
measurements it is very hard to either predict or correct for the
resulting errors. This article outlines an experiment undertaken
to assess the impact on the measurement of reflection coefficient
when using 3.5mm connectors.
Why ?
The ideal connector would not produce any discontinuities in the
transmission line. In reality, as a result of their dimensional
tolerances, there will normally be some small gap between mated
connectors. This gap is often referred to as recession and it may
be both or just one of the mated connectors which are recessed.
The opposite of recession, protrusion, is likely to cause mechanical
damage to 3.5mm devices unless the mating connector is at least
equally recessed and is undesirable. The electrical effect of recession
is to produce a very short section of line having different impedance
characteristics due to the difference in diameter of the center
conductor and the exposed part of the connecting male pin. The theoretical
effect of this change in diameter could be calculated but because
there are many more effects in operation with such a short length
of line, it is simpler to assess practically.
Figure
1-- Connector recession defined

Procedure
The measurements were performed using an HP8722D Vector Network
Analyzer over the range 3GHz to 27GHz at 1601 trace points and with
the bandwidth set to 300Hz. Calibration (network analyzer error
correction) was performed using an HP85052B Precision-grade Calibration
Kit. Three separate calibrations were performed; first with the
sliding load set to zero recession, another with it locked at 2.5
thousandths of an inch (0.0025") recession, and finally with
0.005" recession.
It was important that the characteristics of the item to be measured remained
stable throughout the test, even though the connector's pin recession
was to be varied. The item selected for this task was another 3.5mm
sliding load (HP911D); the type where the connector recession can
be adjusted using a small screwdriver. This enabled the connector
recession to be precisely incremented from zero to 0.005" in
steps of one thousandth of an inch without damaging any of
the components involved in the tests.
Figure
2
Figure 2 forms the reference for the following measurements
and is shown on all plots as the red
line. It was performed with both the calibrating and measured
sliding load set to zero recession. This is considered to be the
correct representation of the voltage reflection coefficient (VRC)
of the tested sliding load.
Figure
3
When the significance of recessed connectors was more generally
recognized, it was suggested that if the test port of the network
analyzer were also recessed and the calibrating and measured devices
had equal recession, then a "good" measurement could still
be performed. Whilst it is no easy task to recess the network analyzer's
test port connector, it is possible to recess the calibration and
measured devices. Figure 3 seems to prove that
as long as the recession is the same during calibration and measurement,
the results are very similar. The small deviations seen (in the
order of 0.0025 VRC) could be accounted for by the unrepeatability
of the connectors and possibly because the reference short and open
circuits were not recessed by the same amount as the sliding load.
Figure
4
Figure 4 shows an ideal calibration performed with no recession
and the measured load adjusted in 0.001" increments from zero
to 0.005". At approximately 22GHz it is apparent that the predictability
of the results changes. As the recession continues to increase,
the VRC reaches a minimum and starts to return towards its original
value.
Figure
5
Figure 5 is similar to Figure 4 but with the errors
"reversed". With 0.005" recession during calibration
then zero recession during measurement, the effect is similar to
that for a connector with protrusion (providing the test port has
sufficient recession to avoid damage). However, the peaks occur
where the troughs were positioned in the previous chart.
Figure
6
In Figure 6, recession during calibration was set midway
between the extremes of recession for measurement. This produced
a plot that resembles an amplitude-modulated signal and clearly
shows the combination of the plots from Figures 4/5.
Conclusions
The results are modified by approximately the same degree but in
the opposite direction depending whether the calibrating, or measured,
sliding load connector was recessed. The small differences are possibly
due to the open and short not being recessed. In addition, the uncertainty
due to use of the connector dial indicator gauge to set the recession
has an effect on measured values.
A Rule-of-Thumb:
- Recession reduces effective directivity
by about 0.005 VRC per 0.001 inch.
- That is equivalent to a typical directivity
of 46dB being reduced to 40dB with 0.001 inch recession,
or to 30dB with 0.005" recession.
|