Viability of Accreditation in Production
About this Article
This article, by Mike Hutchins, was presented at the 1995 NCSL
Workshop & Symposium. At that time the Test and Measurement business
of Agilent Technologies was part of Hewlett-Packard.
Challenges of an ISO/IEC Guide 25 Accreditation
System in a High-volume, Production-oriented Environment
Formal criteria for the assessment and recognition of the technical
element in calibration service provision has existed in some countries
for over a quarter of a century. By contrast, independent appraisal
of quality management practices is a relatively recent development.
Yet the success of ISO9000, as defined by wide-scale pursuit of
registration and its adoption in vendor selection criteria, has
not been mirrored in "accreditation". However, validation
of the product/service actually delivered is beyond the scope of
the standard. For testing (including calibration) this is addressed
by the ISO/IEC Guide 25 / EN45001 standard and it is, perhaps, surprising
that equal emphasis does not seem to have been given by firms to
development of such accredited facilities adequate to meet the total
demand of, the otherwise, quality-conscious industry.
The extent to which the operating requirements of national accreditation
bodies, such as the UK's NAMAS, may have contributed to this situation
is explored from the perspective of the practices of a large, commercial
accredited service organization.
Accreditation in Britain
The National Measurement Accreditation Service was formed in 1985
as a result of merging the British Calibration Service (BCS) with
the National Testing Laboratory Accreditation Scheme (NATLAS) and
is currently a division of the measurement standards bureau, the
National Physical Laboratory. By March 1995 the total number of
NAMAS accreditations exceeded 1700, of which 512 were for calibration.
Established in 1968, the BCS was for many years targeted at high-level
laboratories in government departments and industry. It is not the
oldest such agency -- an honor which belongs to the Australian agency,
NATA, dating back to 1947 -- but has established a respected leadership
role not least because of the illustrated "business results".
Figure 1 shows the number of calibration certificates issued
annually, in total and for the electrical field while Figure
2 illustrates the continuing growth in electrical accreditations.
Recently, the emphasis in this field has been in accreditation for
on-site calibrations although permanent facilities continue to increase
in number too.
Many countries have sought to copy the NAMAS model when establishing
their own accreditation agencies. Its position is supported by a
useful collection of guidance and regulatory documentation, the
creation of which owes much to the contributions of committees including
industry, academia and other government department representatives.
Of particular note in this regard is a recently revised practical
guide to uncertainty determination (1) which also now
addresses non-electrical measurements. The primary criteria for
accreditation (2,3) is based upon ISO Guide 25 with substitution
for subjective language, while further prescriptive detail is provided
by several other documents (4-8) which enlarge upon particular
aspects.
So What's Wrong ?
An indication from a survey made by NAMAS last year supports the
widely-held belief that accredited calibrations account for a fractional
part of the total UK calibration business. This is especially the
case in the electrical field where, even in the accredited laboratories,
NAMAS certifications account only for an estimated 10-20 percent
of calibrations performed. Interestingly, in the dominant mechanical
(dimensional) field, the majority of accredited labs. deliver NAMAS
certification as their standard practice. A possible explanation
for this disparity is discussed later.
Despite NAMAS' published assertion (9) that calibration
integrity is only assured through the service, the majority of customers
seem to be satisfied with company-proprietary calibration certificates.
Although not attesting to the technical validity of the work performed,
most calibration suppliers' quality management systems are ISO9000
registered and this "evidence" of acceptable process control
is typically prominent in the vendor selection criteria of other
ISO9000 compliant companies (over 30 thousand such UK registrations).
This ratio of accredited to non-accredited certifications is certainly
not dissimilar to other countries with recognized accreditation
schemes. The operation of traditional standards laboratories has
provided the basis for the regulations and expectations of assessors.
Whilst untrue for NAMAS, agencies in some countries have preferred
to focus on high-echelon industrial labs., declining requests to
assess general-purpose/lower grade calibration areas or for work
done on customer sites. However, to address the mass-market for
calibration driven by a general trend towards "out-sourcing"
and ISO9000 etc., the overall capability and capacity of the service
represented by accredited facilities must increase.
Considering NAMAS as a typical ISO/IEC Guide 25 agency, which accreditation
criteria or expectations can present implementation difficulty in
a large, production-oriented service operation that may be less
significant to a lower volume standards room? The issues can be
divided into two groups -- the technical, relating to validation
of test methods and logistical, concerning the actual deliverable
that is evident to the customer.
Technical
Probably the single biggest challenge for anyone seeking calibration
accreditation is the requirement to evaluate the measurement uncertainty
of all testpoints in accordance with an internationally recognized
"recipe" such as the ISO Guide to Uncertainty in Measurement,
NIST publication 1297 or, particularly in Britain, NIS3003 (1).
Unfortunately, this scientific technique infers a necessity to
collect multiple data-sets by repeated measurement in order to assess
the type-A or "random" component of the total uncertainty.
The mean for each testpoint data-set would be the actual value reported
in the calibration certificate.
This rigorous approach may be practicable for traditional laboratory
artifacts such as resistors, voltage references or attenuators having
one or two parameters and few points to characterize. However, in
the case of instruments, customers usually expect the calibration
to effectively assess the product for its full capability as indicated
by a specification.
This will often involve measurements at a multitude of points for
different ranges and functions which can, even with a high-degree
of automation, extend to several hours -- days if done manually.
Few customers are willing to pay 3 , 5 or 10 times the cost of a
single calibration that would be asked because of the extra time
required just to perform these repeated measurements. In addition,
longer equipment downtime may be an issue if it's from a production
area. Clearly, a more viable method needs to be devised.
The practice adopted by HP's UK service center and accepted by
NAMAS relies upon a "type-test" assessment carried-out
on the first example of the model encountered. The uncertainties
thus calculated are then attributed to all subsequent calibrations
on that model-type. Figures 3 and 4 present summary extracts
from a paper (10) justifying the method.
|
Figure
4 -- Step-by-step Guide : Allowance for Random Uncertainty
- Assess
and calculate the total systematic (type-B) uncertainty
(Us).
- Divide
the systematic uncertainty by 3. This value is the largest
random (type-A) uncertainty (Ur) that can be tolerated
before it becomes "significant" (>5% of the
quadratic combination used to calculate total uncertainty).
- Carry-out
two complete calibrations of the product and calculate the
absolute differences between data-sets. For each testpoint,
if the difference is found to be "insignificant",
assume Ur to be Us / 3. If random uncertainty
is found to be significant, perform more sets of measurements
at these points and calculate the attributed Ur for
one measurement using:
Ur = Students t x Standard deviation
- where
t relates to the number of repeated measurements
made on the "test-type" sample for 95% confidence.
This formula differs to that generally used by not dividing
sigma by the square-root of the number of measurements as
only ONE measurement is made in subsequent calibrations.
|
There is an acknowledged (unquantifiable) decision-risk concerning
the significance of variation between only two data-sets or whether
more extensive testing is necessary to calculate the real "random"
element. Consequently, consideration should be given to any trends
evident in the series of result-differences; the poorest repeatability
value being attributed to the range/parameter rather than to individual
testpoints. But even using this minimalized method, the time/cost-burden
of performing two calibrations cannot be regimentally followed and
a further short-cut may be sought.
A knowledge of the unit-under-test's operation and construction
can justify assignment of repeatability figures determined for a
similar model/family or even generic category/class, providing the
same procedures and equipment are used. When taking this option,
some attempt at verification at selected testpoints is advisable;
manufacturers sometimes improve their product during its life-cycle
by circuit modification, better component performance, etc., which
may not be apparent by change to the model number. For instance,
earlier "A" versions may not be as stable as the later-released
"B".
To provide confidence that the "type-test" remains valid
for the product population and hence uncertainty assignment in later
calibrations, a sample audit process is necessary which includes
determination of differences between repeated measurements at selected
points and comparison with the "random" allowances. Obviously,
excursion suggests that a full reassessment of the uncertainties
may be required. So far, concentration has been on the "random"
uncertainty -- but what of the type-B or "systematic"
component ?
To meet the requirement of the standard it's necessary to produce
an uncertainty budget which numerically quantifies the possible
error-sources. At HP, like many others, the general practice to-date
has been to base calibration services on the manufacturers recommended
procedures, equipment and testpoints. Owing to the many permutations
that this policy creates, a huge man-power investment is needed
to properly evaluate the measurement accuracies in a manner consistent
with modern-thinking. Having embarked upon this route in 1991 as
part of a strategy to provide "NAMAS calibration" from
our high-production service center (rather than the standards labs.),
documented budgets for several hundred models have been produced.
In volume terms, the top 150 models account for 40 percent of our
30 thousand item calibration business with the remainder falling
into the less than 10 per year category. Continuation with this
model-by-model approach is therefore inappropriate for the thousands
of items where low volumes are expected. This realization has been
supported by a study made by our US colleagues which indicated the
task to require 70 man-years, even for only their existing input.
Of course with new products appearing on the market every day, the
goal would be unattainable with available resources.
Consequently, we are attempting to rationalize our testing methods
and, ideally, develop generic uncertainty budgets based on a more
select test gear list. A downside of this, especially for HP manufactured
instruments, is the potential customer concern that we are deviating
from the recommended performance test as published in a product's
service manual. Another is that existing software developed by various
HP operations worldwide may be unusable for this calibration option;
serious since 70 percent of our input volume is ATE calibrated.
It's clear that with limited resources in any one country, a coherent,
company-wide strategy is needed.
Logistical
Calibration certificates must be in a NAMAS-defined format which
extends to all pages of the report, not simply the first page which
usually carries only item identification, environmental and other
"legal" details which may include a specification status
statement -- more on this later. Measurement data must be supplemented
with uncertainties, set-up conditions and summary description or
reference to the method used.
By tradition, this "full" certificate has become most
customer's expectation but NAMAS does allow an abbreviated or "short-form"
certificate. This is applicable only to equipment having performance
specifications since test results are not presented. An extract
from an example issued for an HP8574 EMI Spectrum Analyzer system
is shown in Figure 5.
Figure 5 -- Extract from a "Short-form"
Certificate for an HP8574A EMC System. The status of 25 tests are
reported altogether on only 6 pages - condensed from about 60 pages
of test data. Click image to enlarge.
It is particularly useful for complex products involving an extensive
calibration routine and where test data is recorded by hand, or
automatically printed in a form which is unacceptable to NAMAS,
so would need to be transcribed. Even with a "short-form"
presentation the certificate design is unique for each model. This
contrasts with the common-look of our single-page "commercial"
certificate, which is rather similar to the cover-page of a NAMAS
certificate described earlier, augmented by a copy of the results
in original format. This practice is common and evidently meets
the needs of the majority of the market; the inference is that accredited
certificate demands are out of step with customer requirements.
For the relatively simple devices often found at standards laboratory
level, it is normal to use calibration data for "corrections"
or in SPC programs. With general-purpose M&TE such use is cautioned
against since complex relationships can exist and insufficient data
is supplied in the calibration for reliable predictions/corrections
to be made. Confirmation, or otherwise, of specification compliance
is the principal need. Most users recognize this and have no interest
in the intricacies of the certified results, preferring a simple
status statement. Its absence means a failure to meet expectations.
However, NAMAS does not require these status remarks to be included
in a "full" certificate. Since "certification"
is often (incorrectly) assumed to imply specification conformity,
there may be a case for recognizing and protecting against misinterpretation
by making status advice mandatory or, where a specification is unknown,
provision of a calibration report rather than certificate.
Where measured values are compared with specifications, NAMAS requires
that acceptance limits are determined by simple arithmetic difference
of specification and uncertainty -- graphically shown in Figure
6.
Although
widely adopted, particularly in the US, NAMAS rejects the producer/consumer
risk management approach so this conservative guardband must be established
whatever test accuracy ratio (TAR) exists. Whilst it assures the customer
that there is very little chance that an out-of-tolerance parameter
is reported as in-tolerance, it can lead to a significant likelihood
that unnecessary adjustment and re-testing will be incurred by the
calibration lab., especially when TARs are low. These costs must be
passed-on to the customer. Further, the desire for a specification
compliance statement cannot be satisfied when measured values fall
into the "gray area". Neither are labs. allowed to quantify
the consumer-risk which might assist the customer meet his own quality
system criteria. Again, selection of accredited certification seems
to make life harder for the purchaser.
Similar to the matter of specification status, NAMAS feels it has
no mandate to define the extent of service provided for any particular
generic equipment-type. Many users are unaware of this, assuming
that the validity of measurements performed (for which the agency
does acknowledge responsibility) also implies "policing"
of the calibration's scope. From the customer's viewpoint it is
another reason to be wary in comparing the offerings from competing
service-providers.
Several documents were published by BCS which attempted to provide
guidance on the extent of calibration for a few instruments but
these were not compulsory. The diversity of modern electronic equipment
makes it difficult to devise a suitable testing procedure applicable
to a generic-type. There are no British Standards prescribing the
performance characteristics and verification of general-purpose
electronic instrumentation which is not used in safety-critical,
consumer or legal metrology situations. In other measurement fields,
such standards exist and are well-established in the user-community.
Thus, "tested in accordance with CISPR 16" (EMC) or "calibrated
against BS 870" (micrometers) concisely describes the extent
of service. These standards often also define the precise equipment
to be used, together with acceptance limits (which allow for uncertainty
in some way accepted in the relevant industry-sector and which supersedes
NAMAS' own policy). Consequently, the ease and commercial viability
with which NAMAS' requirements can be complied with is enhanced.
Having finally produced the calibration certificate, a NAMAS authorized
person must sign it to provide "authentication". This
is intended to be a 100 percent final inspection to confirm correctness
which is manageable for low volumes and comparatively short certificates
issued for traditional laboratory references. Impose that same practice
on multipage certificates for products with many functions/ranges/testpoints
and the task becomes significant and adds delay if a limited number
of authorized personnel are available. Modern quality management
recognizes that "quality" cannot be inspected-in at the
end of the process which seems to be the intent. Acceptance by accreditation
bodies of properly implemented sampling audit (a cornerstone of
most quality management systems) as a means of protecting certificate
integrity rather than signature, would eliminate this associated
cost.
Added Assurance or Caveat Emptor * ?
The foregoing gives an indication why the critical technical assurance
conferred by accredited calibration, in its presently practiced
format, often incurs additional delivery costs. Inevitably, while
these practices are not "standard" in a laboratory, this
extra cost is passed-on to those customers demanding such a service.
This position will become less tenable if a much larger proportion
of the user-community, which is currently indifferent to accredited
calibration because of the cost-premium, alters its purchasing behavior
in appreciation of ISO/IEC Guide 25's purpose and limitations of
ISO9000 registration alone. Commercial pressure to remove any premium
will intensify which, if not effectively policed, may lead to a
reduction in overall product integrity through less extensive testing
as a cost-cutting measure. The "buyer-beware *" caution
would remain and the value of accreditation diminished.
The possibility of this scenario could be averted by developing
a simplified form of accreditation more practically applicable to
general-purpose equipment verification which, while remaining consistent
with the general principles of ISO/IEC Guide 25, allowed more flexibility
in implementation. Such a proposal (11) was produced
last year in order to stimulate debate, especially at national level.
Conclusion
The purpose was to summarize operational aspects of a representative
accreditation program which are felt to be factors in the restricted
provision of accredited calibration, particularly in the electrical
field.
Clearly, a balance must be achieved between metrological ideology
and commercial practice to encourage new or extended accredited
facilities necessary to satisfy the requirements of the mass-market.
Most importantly, such consensus must lead to actual use of these
resources in delivery of all calibrations. The readiness of the
authorities to respond to the sensitivities of today's business
environment will be key to progress, through the co-operation of
the service providers and not by dictate.
If this can be done, it will be to the mutual benefit of assessors,
cal.labs. and equipment users. We will indeed have risen to the
challenge of producing a viable framework to attain the perceived
intent of ISO9000 - real "quality" in the product, not
just the process.
References
- NAMAS NIS3003 (1995, Edition 8) - Uncertainty and Confidence
in Measurement
- NAMAS M10 & Supplement - General criteria of Competence
- NAMAS M11 - Regulations
- NAMAS M18 - Accreditation for Site Calibration & Testing
- NAMAS M13 - Conditions for Use of NAMAS Logo [includes cal label
design spec.]
- NAMAS M25 - Certificates of Calibration
- NAMAS M16 - Quality Manual Preparation Guide
- NAMAS M51 - Quality Audit & Quality System Review
- Rogers, J. (NAMAS), "Validity of Calibration & Test
Data" : Paper published by NAMAS contrasting ISO9000 with
EN-45001 and summarized in their Newsletter (Spring 1994).
- Instone, I. (Hewlett-Packard Ltd., Winnersh, UK), "Calculating
the Uncertainty of a Single Measurement" : Institution of
Electrical Engineers (London), digest number 1993/109.
- Abell, D. (Hewlett-Packard Co., Mountain View, Ca.), "Proposal
for Simplified M&TE Accreditation", February 1994: Public
discussion paper drawing upon ideas and work of the NCSL TQM Committee
on Calibration System Requirements in relation to ANSI/NCSL Z-540-1
development and submitted to senior management of organizations
including NAMAS, NIST, NAVLAP, Standards Council of Canada and
EAL.
|