Integral Crystal Oscillator Testing
Introduction
A policy concerning the testing of the reference
oscillators contained within many instruments such as counters,
synthesizers and spectrum analyzers has been defined and adopted
in Agilent's service centers worldwide.
Quartz crystal oscillators are used extensively
in measurement instruments to provide an accurate reference frequency
or timebase against which measured signals may be compared.
There are two characteristics of such devices
which users may wish to know:
- The absolute oscillator frequency once
it has warmed up
- The stability of the oscillator frequency
over time.
Our standard policy is to test the first
parameter (and to adjust where necessary) but not the second. This
article is designed to help explain the rationale for this decision.
High-stability oscillators such as the HP10811A/B
have their stability specified in terms of a drift rate per day.
This means that within a 24-hour period the frequency will not change
by more than the permitted tolerance over that period. This is in
contrast to lower stability oscillators whose aging rate is stated
over a month or a year. The frequency may well change in the short
term more than the specified amount, but the long-term trend will
comply with the specification -- see Figure 1.
 |
| Figure 1-- A representation
of the frequency stability of an uncompensated or temperature
compensated quartz crystal oscillator. Note how the aging rate
would far exceed, or be be far less than, specification if measurements
were taken over a shorter period than the specified one month
for this oscillator. The apparent aging rate would also vary
considerably depending when the measurements were made. |
Experience
Our company has been manufacturing precision crystal oscillators
for many years (formerly as Hewlett-Packard), and during that time
extensive analysis of their performance has been made. In fact,
hundreds of thousands of high-stability oscillators have been sold
by Agilent/HP, which gives a significant population from which to
base our strategy for testing the devices.
The analysis has determined that it is not generally necessary
to test the drift rate of high-stability oscillators. In fact, it
has been estimated that there is a 99.995% probability that a 10811-series
oscillator that performs to its other specifications will also pass
the aging rate specification. This assumption has been carried across
to similar products not manufactured by us. It is also a fact that
an aging-rate check is seldom part of the recommended calibration
procedure for any instruments fitted with precision crystal oscillators,
whoever the manufacturer.
Test Procedure
|
For confidence and repeatability, the oscillator is allowed
to warm up according to the following minimum times:
- Room temperature and temperature compensated crystals:
1 hour
- Medium performance oven: 12 hours
- High performance oven: 48 hours
Once the instrument has warmed up, its absolute frequency
is measured. This is compared to test limits that have been
established for each oscillator type and, if necessary, the
frequency is optimized to nominal.
|
 |
Drift Testing
Drift testing is only undertaken for devices with a 24-hour specification.
For all other types, stability testing would require at least 30
days for a specification compliance statement to be made. Such measures
would simply not be worthwhile. An illustration of the inadequacy
of short-term drift testing is shown in Figure 1 where changes
observed over any 24-hour period could potentially give a very misleading
indication as to the oscillator's long-term performance.
Supplementary Test
To satisfy the needs of customers who want confirmation that their
High Performance Oven Oscillators meet their drift specification,
the aging rate measurement can be performed as a supplementary test
at an additional charge. The oscillator is monitored constantly
over successive 24-hour periods until compliant stability is shown.
For some devices this may take up to ten days to occur (greater
than ten days would be taken as a failure). The final value is then
recorded as an offset from the nominal value, the drift over the
24-hours recorded and presented both numerically and graphically
-- see Figure 2.
 |
| Figure 2-- A representation
of the warm-up characteristic of a high stability, ovenized
quartz crystal oscillator. In this example the drift or aging
rate specification is 1 part in 109 per day. |
Summary
- For all instruments fitted with an internal reference oscillator
we will report on the absolute reference frequency and will adjust
to nominal when it is outside of predetermined limits for that
particular oscillator type.
- Aging-rate (drift or long-term stability) measurement will be
undertaken as a supplementary chargeable service, on request,
for oscillators with a daily aging specification. This may add
as much as ten days to the instrument calibration turn-around
time.
Note: For instruments that are
referenced to an external frequency standard such as a Cesium Beam
or Rubidium Vapor device, determination of an integral oscillator's
aging rate has no value to the user.
|